Tek 2465B calibration issue
George Kerber
I have a 2465B that passes all calibration routines except parametric
measurements CAL 09 steps 2 and 3 which always indicate "LIMIT". All of the
rest of the CAL 09 steps pass. CAL routines 01, 02, 03 have been been
repeated a couple of times with no problems. The scope functions fine and
seems to perform parametric measurements correctly.
Has anyone seen this problem before and have a fix? Any thoughts would be
appreciated.
George Kerber
www.glkinst.com
TekMan
appreciated.
George Kerber
www.glkinst.com
download the servicemanual of the 2465B from
www.teknetelectronics.com
and follow the cal. procedure. This helped with my tek2465B
hth,
Andreas
George Kerber
Andreas,
That is exactly what I did. The entire cal procedure has been perfomed a
couple of times. Still get "LIMIT" on CAL 09 steps 2 & 3. Everything else
passes and the scope appears to function normally, except at startup it
fails test 04 code 02 which means one or more cal coefficients is out of
limits. I assume this is from CAL 09 steps 2 & 3.
George
TekMan
George Kerber said:
Andreas,
George
...snippety..
George,
okay, when you can't complete by simply caloibrating:
this says that your scope can't be cal'd any more due to a hardware
problem on the board (most often input amp chips partly defective/out
of specs).
change these chips. But be warned: Spares are ahrd to get.
hth,
Andreas
Jim Yanik
[email protected] (TekMan) wrote in
..snippety..
George,
of specs).
hth,
Andreas
Isn't there something in the service manual that relates the cal step to a
particular circuit area? ISTR that there was.
0
836
Andrew Holme
0
1K
joeboatertx
8
4K
Jim Yanik