What do you mean by delta T. Do you wish to measure the temperature difference between two position or the temperature change at one position.
To measure the temperature difference of two positions, put a junction in each position and balance one voltage against the other.
To measure a temperature change, you will need a cold junction reference and some form of recording, pencil and paper, computer etc.
You can look up the sensitivity of a T type thermocouple and calculate the voltage you will get. For under 1K measurements, you will need a very accurate measurement system with low drift.
Edit The sensitivity is 39µV/K so 0.35K will give only 13.7µV.
Thank you for your reply. Some more questions before I actually start making the changes in my experimental set up.
Please correct me if I am wrong, I have been doing a lot of googling through the night and this is my take,
I understand that the two junctions will be at the two points,(like a M shape); the net result of this will be to read a voltage which would correspond to the differential temperature.
Now the wires that come out can no longer be connected to the data logger as temperature measurement because the two ends would be of the same material Cu, in the case of T type.
Also, we don't want the data logger's inbuilt algorithm to do any cold end compensation.
Thus the wires would be connected to the data logger as a voltage reading.
How do I proceed from here? Should I collect the voltage data and put it into an excel file and use the eight order polynomial to convert the voltage to temperature? And what would the accuracy of this reading be? What is the general method to arrive at the accuracy of such measurements?
Thanks in advance
Kamal