Only an old fashion cap checker that checks for leakage at 600 vdc will work. An ESR meter (I have one), will not find this problem, besides they only work on values of 1 Ufd or above. Ed.You must also replace 19C68 (.01 @ 630 v). They are supposed to be rated at 630 but are marked 600v. This is a common problem in this model scope. Those caps develop leaks at voltages above 200. I just fixed 2 of them. High voltage to this crt anode is 20k, higher than any scope I have used. This is why this scope has a better defined focus than any other but comes at a price as the years go by. Ed.
-