I have a transformer here I believe is malfunctioning. I wanted a second opinion and a second pair of eyes on this datasheet to make sure. The transformer is custom made but the company sent me the data sheet. a file has been uploaded of the datasheet. Values I found during inspection below.
leads at wht/blu and brown = 120v
leads at blue and wht/brn = 120v
leads at blk and red = 0v
leads at white and white = 19v
leads at yellow and orange = 0v
leads and gray and violet = 0v
With these numbers, I'm thinking the transformer needs replaced. Can someone verify or deny this theory with reasons as to why? The part that i'm unsure on is that white is a self-reset thermal switch and I'm unsure as to whether I should see 19v there. Other than that, I'm seeing voltage in but the outputs don't have anything coming out. Your responses are always much appreciated.
leads at wht/blu and brown = 120v
leads at blue and wht/brn = 120v
leads at blk and red = 0v
leads at white and white = 19v
leads at yellow and orange = 0v
leads and gray and violet = 0v
With these numbers, I'm thinking the transformer needs replaced. Can someone verify or deny this theory with reasons as to why? The part that i'm unsure on is that white is a self-reset thermal switch and I'm unsure as to whether I should see 19v there. Other than that, I'm seeing voltage in but the outputs don't have anything coming out. Your responses are always much appreciated.