no i did not..
ad620 is an instrumentation amplifier so 741 is not for this work.. 741 uses for just increasing gain because ins. amp. gain is not enough to see all waveforms of PQRST..
Could you explain this in more detail? The instrumentation amplifier draws a negligible input current (max. 4nA over the full temperature range). Therefore the losses in the cables leading to the human subject should be negligible, too (e.g. even at 100 Ohm the voltage drop will be <= 400nV, much less than the millivolts you expect to measure).i used copper cable and -as u can guess- loss is too much
thanks for your interest and advices