Hello Howard,
Thanks for your interesting message!. I will try that ASAP.
But perhaps I can help you on this:
I had some information stored in my PC for 'future use'... and here is
the future
!
I hope Jim W8ZR doesn't care for forwarding this information, as he
sent it to help other people with similar troubles.
From the TekScopes Yahoogroup:
=================
Gang,
I thought I'd pass on some notes on calibrating my "new" 2465B, which
is now freshly aligned and running perfectly. I owe Denis Cobley a
huge debt of gratitude for walking me through some stumbling blocks.
Hope these notes can spare some other owners some of the frustration I
experienced.
Here is the problem I experienced:
Before calibration, I was getting an ERROR 04 02 message at powerup,
signifiying that one or more stored calibration constants was out of
spec.
Using a time mark and calibration generator, I stepped throught CAL
01 (horizontal calibration) and CAL 02 (vertical calibration). At a
few of the steps, I got a LIMIT error, and had to back up a step and
redo it..
After I finished, I powered up the scope but still got the ERROR 04
02 message. I redid the entire calibration, but still couldn't get
rid of the error message.
Also, the display showed a row of dots along the bottom line of the
readout, where spaces were supposed to be. WHen I moved the
calibration jumper back to its normal position, the dots turned into
question marks.
Here's what I learned. The service manual doesn't say this, but none
of the calibration constants are stored until the entire calibration
routine (e.g., CAL 01, CAL 02, etc.) is completed and exited. In other
words, if only some of the steps are completed and then the user exits
the calibration routine, none of the new constants are saved..
Second, if a LIMIT error is encountered somewhere during the
calibration, then the new constants won't be stored. The only solution
is to exit the routine and start over.
Finallly (and this is what was causing my problem), there are three
undocumented steps in the CAL 01 procedure. The service manual shows
31 steps for CAL 01, but the 2465B actually has 34 steps. After
completeing the 31 steps, it is necessary to advance through 32, 33,
and 34, and then let the scope exit the routine by itself. If you
don't do this, none of the calibration constants will be stored.
After completing the entire calibration in this way, the dots and
question marks disappeared, as did the startup error message, and all
is as it should be.
Regards,
Jim Garland W8ZR
===============
Thanks again, Jim!
Best regards,
JOSEhttp://jvgavila.com