Newbie-to-Tek-scope-repair here. I was given a 465 for free. I would like to repair it.
Initial symptoms as reported by previous owner: no activity on panel upon startup, fan only.
Initial power-off inspection and investigation: nothing showed on visual (no dry soldering, no burnt or otherwise iffy-looking components). Didn't smell anything that was suspect. Continuity all good with fuses, plug to switch activation, etc.
Power-on investigation: all rails showed good except the 5v, it tested -0.45. Re-examination of 5v area using a 30x loupe showed a cracked rectifier (no other abnormalty, such as burning, etc). Pushing on one side with a dental tool caused the halves to shift slightly.
I replaced the rectifier, double-checked solder joints, and powered up. The rectifier unsoldered itself within 3 sec. I removed the rectifier and re-checked the other rails, all good.
So, what likely happened?