Hey there, im new here so first of all, hi everybody!
First of all, i dont know much about circuits or electronics, only the basics of the basics so i need your help...
So, i got an old Yugoslav military radiation detector. The unit was produced in 03/1986 and when i put the batteries in it works, but i can see that it is struggling to give out accurate readings.
In the BAT bracket (battery powered 2 x 1,5V D) :
-KN - Battery control
-cGy/h - upper scale for use with the ionization chamber
-μGy/h - lower scale for use with the Geiger Mühler tube
-ISKLJ. - off position
In the AKU. bracket (powered from 12/24V car battery through an adapter)
-cGy/h - upper scale for use with the ionization chamber
-μGy/h - lower scale for use with the Geiger Mühler tube
The problem is while running on 1.5V batteries the battery control works fine (KN) and i tested the Geiger - Mühler tube on a radioactive source in the μGy/h position and it gives out accurate readings, the problem is the ionization chamber. when i try to calibrate with the KAL marking on the scale, the needle doesn't drop to 0 but stays at 0.5 (not 0.05), and i have also seen that when i leave the device turned on for about 10 minutes and i set it back to the cGy/h position, i get slight variations on the needle (sometimes it shows 0.5 sometimes 0.4)
here you can see that the device is working fine on the mikro-gray area (as seen in the video i tested it on a radiation source), but while operating in the high-range area (cGy/h) with the ionization chamber, when the device is calibrated as per instructions, the needle won't drop below 0,4-0,5 cGy/h - and it should drop to zero.
Now, there are 2 calibration blocks inside the device with 7 or 8 potentiometers that are used to factory-calibrate the device. But i don't know which one is used for what... I have a schematic of the device but sadly it is not the original one. It's from a Czechoslovak copy of the device called Intenzimetr IT-65.
The layout of the circuits is the same, but the Czechs used their own domestically produced elements inside. This device also uses old capacitors which probably need replacement, but if any element inside the device is changed, it will need to be re calibrated.
Schematic of the IT-65, the Czechoslovak copy of the DR-M3. The Yugoslav version has different circuits on the calibration panels and the power supply panel , so this schematic can only be used to see the layout of the fixed inner circuits of the chassis:
Here i also have some pics of the insides.
This is the power supply block:
The circuit elements on the uGy/h callibration block:
Underside of the ionization chamber:
What is your opinion on what could be causing the problem, that the needle doesn't drop to zero, but stays at 0,4-0,5 when the device is calibrated?
First of all, i dont know much about circuits or electronics, only the basics of the basics so i need your help...
So, i got an old Yugoslav military radiation detector. The unit was produced in 03/1986 and when i put the batteries in it works, but i can see that it is struggling to give out accurate readings.
In the BAT bracket (battery powered 2 x 1,5V D) :
-KN - Battery control
-cGy/h - upper scale for use with the ionization chamber
-μGy/h - lower scale for use with the Geiger Mühler tube
-ISKLJ. - off position
In the AKU. bracket (powered from 12/24V car battery through an adapter)
-cGy/h - upper scale for use with the ionization chamber
-μGy/h - lower scale for use with the Geiger Mühler tube
The problem is while running on 1.5V batteries the battery control works fine (KN) and i tested the Geiger - Mühler tube on a radioactive source in the μGy/h position and it gives out accurate readings, the problem is the ionization chamber. when i try to calibrate with the KAL marking on the scale, the needle doesn't drop to 0 but stays at 0.5 (not 0.05), and i have also seen that when i leave the device turned on for about 10 minutes and i set it back to the cGy/h position, i get slight variations on the needle (sometimes it shows 0.5 sometimes 0.4)
here you can see that the device is working fine on the mikro-gray area (as seen in the video i tested it on a radiation source), but while operating in the high-range area (cGy/h) with the ionization chamber, when the device is calibrated as per instructions, the needle won't drop below 0,4-0,5 cGy/h - and it should drop to zero.
Now, there are 2 calibration blocks inside the device with 7 or 8 potentiometers that are used to factory-calibrate the device. But i don't know which one is used for what... I have a schematic of the device but sadly it is not the original one. It's from a Czechoslovak copy of the device called Intenzimetr IT-65.
The layout of the circuits is the same, but the Czechs used their own domestically produced elements inside. This device also uses old capacitors which probably need replacement, but if any element inside the device is changed, it will need to be re calibrated.
Schematic of the IT-65, the Czechoslovak copy of the DR-M3. The Yugoslav version has different circuits on the calibration panels and the power supply panel , so this schematic can only be used to see the layout of the fixed inner circuits of the chassis:
Here i also have some pics of the insides.
This is the power supply block:
The circuit elements on the uGy/h callibration block:
Underside of the ionization chamber:
What is your opinion on what could be causing the problem, that the needle doesn't drop to zero, but stays at 0,4-0,5 when the device is calibrated?