Recall - Fluke 33X Clamp Meters

A

aussiblu

see http://www.recalls.gov.au/content/index.phtml/itemId/964924

Fluke Australia Pty Ltd-Fluke 33X Clamp Meters
PRA number: 2009/10778
Date created: 4th May 2009
Product information
Product description
Fluke 333, 334, 335, 336 and 337 Digital Clamp Meters
manufactured between 3 March 2008 and 27 February 2009.
Identifying features
Fluke 333, 334, 335, 336 and 337 Digital Clamp Meters
What are the hazards?
Electrocution due to incorrect reading
What are the defects?
A defect has the potential, over time, to cause the meter to
display a low or no voltage reading on a circuit energised with
a hazardous voltage. In this situation it is imperative that the
practice of testing the meter on a known live circuit before and
after checking the circuit under test is maintained. Should the
meter not pass the "known live circuit test" its use should be
discontinued immediately, identified as faulty and removed from
service.
Where the product was sold
a.. National
Supplier
Fluke Australia Pty Ltd
Supplier's web site
http://www.fluke.com.au
What should consumers do?
Cease using and immediately return suspect products to the place
of purchase to be replaced free of charge. For further
information visit www.fluke.com.au or phone (02) 8850 3333.
Advertisements and supporting documentation
 
A

aussiblu

and this one:
http://www.recalls.gov.au/content/index.phtml/itemId/955471

Fluke-T2 Electrical Tester
PRA number: 1998/3706
Date created: 12th November 1998
Product information
Product description
Serial Numbers Less Than 70524051. Units That Have An "r"
Stamped After The Serial Number Have Been Repaired And Are No
Longer Affected. Sold In Australia Since February 1998.
What are the hazards?
Electric Shock
What are the defects?
A Battery Contact To The Circuit Board Inside The Instrument Can
Corrode Over Time Causing It To Operate Intermittently. When The
Malfunction Occurs The Meter May Not Indicate That Voltage Is
Present Placing The User In Potentially Hazardous Situation.
Where the product was sold
a.. National
Supplier
Fluke Australia Pty Ltd
What should consumers do?
For Information Ring Philips Business Electronics On (02) 9805
4486. Units Will Be Serviced And Returned Directly To Customers
At No Charge.
 