Sir Gopakumar . . . . .
I just hope that this unit did not end up coming from, and being stored for decades in an old MAERSK shipping container in Bhubaneswar.
Where its
HOT-HOT-HOT and 118% humidity . . . with the added extra . . . from not having mere beads of sweat . . . but a continuous sheet of perspiration being on oneself continually.
That then relates to the possible condition of the myriads of slide switchwes used on the units front panel.
In that time frame of build, and on this quality of unit, they typically used SILVER flashed contacts . . . and in those climatic conditions, could leave all BLACK Ag O
2 on those contacts.
Next, direct your attention to the
YELLOW mark up section . . . . and there . . . . your VERY first consideration, would be the
RED markup of the INT-EXT switch, in seeing for sure, that it is being in the
INT position. If not . . . . . positively no side to side Hoz sweep.
Next, would be its having bad internal connectivity, which I just referred to, so flip it up and down for about 25 times, to hope for some frictional cleaning action, for further evaluation .
The next item, would be the MAG switch and its like markup . . . also . . . . being just above .
Poor connectivity on it can also cause a loss of Hoz sweep.
FACEPLATE REFERENCING . . . . .
73's de Edd
.....