Hi all,
I have some questions about the manual of Agilent 33250A,
"
Ground Loops
Except for its remote interface connectors and trigger connector,
the 33250A is isolated from chassis (earth) ground. This isolation helps
to eliminate ground loops in your system and also allows you to reference
the output signal to voltages other than ground. The illustration on the
following page shows the function generator connected to a load through
a coaxial cable. Any difference in ground potentials (VGND) will tend to
make current IGND flow in the shield of the cable, thus causing a voltage
drop due to the shield’s impedance (ZShield). The resulting voltage drop
(IGND x ZShield) appears as an error in the load voltage. However, since
the instrument is isolated, there is a high series impedance (typically
1 MΩ in parallel 45 nF) in the path to oppose the flow of IGND and
thereby minimize this effect.
"
My questions are:
1. How does the Vgnd being produced? If there is a Ignd, then it means chassis ground and earth ground are connected to each other(which is used as the common method to prevent shock hazard), then shouldn't the potential difference be zero?
2. May I say that even if there is no Ignd, the VL is still not equal to Vout due to the exist of Zshield?
3. Is the source signal on earth a floating one or grounded one?
I have some questions about the manual of Agilent 33250A,
"
Ground Loops
Except for its remote interface connectors and trigger connector,
the 33250A is isolated from chassis (earth) ground. This isolation helps
to eliminate ground loops in your system and also allows you to reference
the output signal to voltages other than ground. The illustration on the
following page shows the function generator connected to a load through
a coaxial cable. Any difference in ground potentials (VGND) will tend to
make current IGND flow in the shield of the cable, thus causing a voltage
drop due to the shield’s impedance (ZShield). The resulting voltage drop
(IGND x ZShield) appears as an error in the load voltage. However, since
the instrument is isolated, there is a high series impedance (typically
1 MΩ in parallel 45 nF) in the path to oppose the flow of IGND and
thereby minimize this effect.
"
My questions are:
1. How does the Vgnd being produced? If there is a Ignd, then it means chassis ground and earth ground are connected to each other(which is used as the common method to prevent shock hazard), then shouldn't the potential difference be zero?
2. May I say that even if there is no Ignd, the VL is still not equal to Vout due to the exist of Zshield?
3. Is the source signal on earth a floating one or grounded one?