  • Thread starter Anthony Fremont
  • Start date
A

Anthony Fremont

Got my new Extech MM560 today, pretty cool. Haven't had much opportunity to
play with it much yet, but I did a few measurements just to see how the 15
year old Micronta was holding up.

On an old 6V battery I use for tinkering I get this.
Extech: 5.833V
Micronta: 5.98V (2.5% high)

Plugged into the wall:
Extech: 123.61V
Micronta: 124.9V (1% high)

Running PIC Current:
Extech: 1.387mA
Micronta: 1.40mA (<1% high)

Sorry, no resistance or really low current measurements yet. I do see that
the power company is running the mains at anywhere from 59.9xxHz to 60.0xxHz
though constantly sliding around. :)

Concerns: Build quality seems decent enough. The probes are certainly
nicer than the Micronta, but we'll have to see how they look after 15 years.
;-) The meter looks good so far, but it is different from what I'm used to.
When you first set to a uA or mA mode, there is a residual charge of some
sort that counts down for a minute and then settles down to below a count of
8-10 (starts out around 400). Shorting the probes makes no difference, you
just have to give it a minute to settle down. Is that normal?

Does anyone else think it odd that the battery was already installed?
 
B

Ben Jackson

When you first set to a uA or mA mode, there is a residual charge of some
sort that counts down for a minute and then settles down to below a count of
8-10 (starts out around 400). Shorting the probes makes no difference, you
just have to give it a minute to settle down. Is that normal?

Thermal converter for the True RMS? :)
 
A

Anthony Fremont

Ben said:
Thermal converter for the True RMS? :)

I called Extech about it, they said they didn't think that the meter should
be doing that, especially since shorting the probes makes no difference, but
the fuse is fine. I'm taking their advice and contacting Mouser about it to
see what they want to do. After another five or ten minutes, the count
hovers around 4. It really does act like something is heating up, except
that turning it off and right back on starts the process over. Strange how
it's completely independant of any measuring you're doing. I guess it's a
cap/voltage/charge thing affecting some ADC reference chip or something. I
just emailed Mouser's tech suport arena, we'll see what they say.
Disappointing to say the least. :-(

If everybody wanted to be really gentelmanly now, they could send me a meter
with an NIST cert. ;-)
 
A

Anthony Fremont

Anthony said:
If everybody wanted to be really gentelmanly now, they could send me
a meter with an NIST cert. ;-)

Everybody, of course meaning Extech and Mouser. :)
 
E

ehsjr

Anthony said:
I called Extech about it, they said they didn't think that the meter should
be doing that, especially since shorting the probes makes no difference, but
the fuse is fine. I'm taking their advice and contacting Mouser about it to
see what they want to do. After another five or ten minutes, the count
hovers around 4. It really does act like something is heating up, except
that turning it off and right back on starts the process over. Strange how
it's completely independant of any measuring you're doing. I guess it's a
cap/voltage/charge thing affecting some ADC reference chip or something. I
just emailed Mouser's tech suport arena, we'll see what they say.
Disappointing to say the least. :-(

If everybody wanted to be really gentelmanly now, they could send me a meter
with an NIST cert. ;-)

Sorry to hear it - what a PITA.

Maybe they each should end you one of each (w/without the
NIST) so you can compare and see how they should behave. :)

That's a nice looking meter from the specs - thanks for
posting the url in the other thread. It has a very large
"drool factor". Somehow I couldn't find that on the
spec sheet! :)

Ed
 
A

Anthony Fremont

ehsjr said:
Sorry to hear it - what a PITA.

It'll be a hassle sending it back, but oh well. It still seems to work fine
outside of that little issue. I'm most surprised that my old meter has held
up as well as it has.
Maybe they each should end you one of each (w/without the
NIST) so you can compare and see how they should behave. :)

I figured they could at least send me a certified one so we could all feel
good knowing it would be right this time?
;-)
That's a nice looking meter from the specs - thanks for
posting the url in the other thread. It has a very large
"drool factor". Somehow I couldn't find that on the
spec sheet! :)

It's not quite as flashy as a bright yellow Fluke or a firey red Wavetek,
but it ain't bad. Decent specs and for the money it seemed the best deal.
 
M

MassiveProng

It'll be a hassle sending it back, but oh well. It still seems to work fine
outside of that little issue. I'm most surprised that my old meter has held
up as well as it has.


I figured they could at least send me a certified one so we could all feel
good knowing it would be right this time?
;-)


It's not quite as flashy as a bright yellow Fluke or a firey red Wavetek,
but it ain't bad. Decent specs and for the money it seemed the best deal.


Should have spent $150 and bought the used HP, which is likely to be
an order of magnitude more accurate, and more reliable, and less
likely to drift over time.

PERFECT for long term home use without the need for a yearly cal
session!

Bwuahahahahahahaha!
 
M

MassiveProng

Thermal converter for the True RMS? :)


He is probably using the 4 to 20 mA scale which is NOT for current
measurement, but for process instrument measurement.

It RESTS at 4! & EXPECTS 4 as a baseline Minimum! D'oh!

Took me all of 6 seconds looking at the data sheet.
 
A

Anthony Fremont

MassiveProng said:
He is probably using the 4 to 20 mA scale which is NOT for current
measurement, but for process instrument measurement.

It RESTS at 4! & EXPECTS 4 as a baseline Minimum! D'oh!

Took me all of 6 seconds looking at the data sheet.

I know how to take current measurements, and I know how to use a meter in
general. Reread the post til you understand that the count starts at 400ish
(4.xx mA in mA mode, 4.xx uA in uA mode) and drifts down to around 4 (0.04mA
or 0.04uA) and that shorting the probes has no effect at all. It has
nothing to do with process measurements or current loops either. If you'd
have spent another 6 seconds, you would have realized that process current
loop measurements are displayed as percentages, not ordinary mA or uA.

I talked to Extech about it and they think that it's a defective meter.
They said to send it back.
 
D

David L. Jones

Got my new Extech MM560 today, pretty cool. Haven't had much opportunity to
play with it much yet, but I did a few measurements just to see how the 15
year old Micronta was holding up.

On an old 6V battery I use for tinkering I get this.
Extech: 5.833V
Micronta: 5.98V (2.5% high)

Plugged into the wall:
Extech: 123.61V
Micronta: 124.9V (1% high)

Running PIC Current:
Extech: 1.387mA
Micronta: 1.40mA (<1% high)

Sorry, no resistance or really low current measurements yet. I do see that
the power company is running the mains at anywhere from 59.9xxHz to 60.0xxHz
though constantly sliding around. :)

Concerns: Build quality seems decent enough. The probes are certainly
nicer than the Micronta, but we'll have to see how they look after 15 years.
;-) The meter looks good so far, but it is different from what I'm used to.
When you first set to a uA or mA mode, there is a residual charge of some
sort that counts down for a minute and then settles down to below a count of
8-10 (starts out around 400). Shorting the probes makes no difference, you
just have to give it a minute to settle down. Is that normal?

DC or AC range?

The lower DC ranges on higher spec meters often have Gohm input
impedance which pick up noise, but goes away instantly when you short
the probe.

Dave :)
 
A

Anthony Fremont

MassiveProng said:
Should have spent $150 and bought the used HP, which is likely to be
an order of magnitude more accurate, and more reliable, and less
likely to drift over time.

PERFECT for long term home use without the need for a yearly cal
session!

Bwuahahahahahahaha!

How many times do we have to go thru this? I need/want a portable meter,
not a bench meter. I need/want .01uA resolution, not 1uA resolution. Other
than that, I'm sure it's a fine meter. That's why the 3478a is not the only
thing made.
 
D

David L. Jones

Should have spent $150 and bought the used HP, which is likely to be
an order of magnitude more accurate, and more reliable, and less
likely to drift over time.

PERFECT for long term home use without the need for a yearly cal
session!

You don't get it do you? People buy a hand held meter for many good
reasons, a big bench meter is not an option for them.

Dave :)
 
A

Anthony Fremont

David said:
DC or AC range?

The lower DC ranges on higher spec meters often have Gohm input
impedance which pick up noise, but goes away instantly when you short
the probe.

Shorting the probes has no effect. :-( Oddly enough, it only does then
when you have AC+DC (which was apparently the default) or just AC selected.
If you select DC only then it seems fine. OTOH, if you select only AC
current, it does the countdown thing, only slightly differently (it takes
the same amount of time to get down near zero, it just updates the display
several times/second instead of once every couple of seconds). Strange. I
wish someone else had one and could tell me there experience.
 
D

David L. Jones

Shorting the probes has no effect. :-( Oddly enough, it only does then
when you have AC+DC (which was apparently the default) or just AC selected.
If you select DC only then it seems fine. OTOH, if you select only AC
current, it does the countdown thing, only slightly differently (it takes
the same amount of time to get down near zero, it just updates the display
several times/second instead of once every couple of seconds). Strange. I
wish someone else had one and could tell me there experience.

Sorry, thought you were talking about the voltage range :-/

My Fluke 87III does the same thing on the uA and mA AC current range.

In the 4.5 digit uA mode it starts out out around 2.8uA and slowly
settles down over about 30 seconds to around 0.18uA and won't go
lower.
In 3.5 digit mode it starts around 3.5uA and drops to around the same
0.1-0.2uA, but it's faster than the 4.5 digit mode.
mA range is similar with a residual offset of around 1.8mA

Obviously the True RMS converter has some settling time, and a
residual offset which you can null out, but it takes time again to
settle down to zero after the null.

DC is of course fine as there is no True RMS converter to deal with.

I don't think there is anything wrong with your meter, it's fairly
normal for True RMS meters.

Meters without True RMS don't exhibit this problem, they just go to
zero straight away and sit there.

Dave :)
 
D

David L. Jones

He is probably using the 4 to 20 mA scale which is NOT for current
measurement, but for process instrument measurement.

It RESTS at 4! & EXPECTS 4 as a baseline Minimum! D'oh!

Took me all of 6 seconds looking at the data sheet.

No, it's the True RMS converter.

Dave :)
 
D

David L. Jones

He is probably using the 4 to 20 mA scale which is NOT for current
measurement, but for process instrument measurement.

It RESTS at 4! & EXPECTS 4 as a baseline Minimum! D'oh!

Took me all of 6 seconds looking at the data sheet.

You don't know your own equipment again! - check your beloved Protek
506, it does a similar thing. Takes almost 30 seconds to settle down
to zero on AC current, but it does get to zero, no apparent offset.

Dave :)
 
A

Anthony Fremont

David said:
Sorry, thought you were talking about the voltage range :-/

My Fluke 87III does the same thing on the uA and mA AC current range.
Hmm......

In the 4.5 digit uA mode it starts out out around 2.8uA and slowly
settles down over about 30 seconds to around 0.18uA and won't go
lower.
In 3.5 digit mode it starts around 3.5uA and drops to around the same
0.1-0.2uA, but it's faster than the 4.5 digit mode.
mA range is similar with a residual offset of around 1.8mA

Obviously the True RMS converter has some settling time, and a
residual offset which you can null out, but it takes time again to
settle down to zero after the null.

I wonder why the tech support guy didn't snap to that? I guess he probably
figured I had it in straight DC mode.
DC is of course fine as there is no True RMS converter to deal with.

I don't think there is anything wrong with your meter, it's fairly
normal for True RMS meters.

Meters without True RMS don't exhibit this problem, they just go to
zero straight away and sit there.

Ok then, I may keep it for a while and see how it does. It certainly seems
accurate. It has a 3yr warranty, and it certainly seems like Extech is
willing to stand behind their product. I just did some tests and sure
enough in DCV mode if I press SELECT and chose RMS, it does the same kind of
thing. It takes almost two minutes to really settle out. I don't think I
like that, but it sounds like it is par for the course. Really updates the
display slowly doing RMS measurements too (like once every two seconds).

I just did a couple of resistance measurements:

1M resistor
Extech: 1.0195M
Micronta 1.020M (.05% high)

390R
Extech: 394.44R
Micronta: .395K (.14% high)

I don't know, but I think my Micronta is not too shabby.
 
D

David L. Jones

I wonder why the tech support guy didn't snap to that? I guess he probably
figured I had it in straight DC mode.

Yeah, or they might simply follow the "customer is always right"
policy, and just wanted to keep you happy.
Ok then, I may keep it for a while and see how it does. It certainly seems
accurate. It has a 3yr warranty, and it certainly seems like Extech is
willing to stand behind their product. I just did some tests and sure
enough in DCV mode if I press SELECT and chose RMS, it does the same kind of
thing. It takes almost two minutes to really settle out. I don't think I
like that, but it sounds like it is par for the course. Really updates the
display slowly doing RMS measurements too (like once every two seconds).

I'm not really a huge fan of True RMS meters, they are slow and
generally not as accurate as normal meters. If you *know* you have a
sinusoidal signal then a normal meter is better.
Switchable true RMS would be nice!
I just did a couple of resistance measurements:

1M resistor
Extech: 1.0195M
Micronta 1.020M (.05% high)

390R
Extech: 394.44R
Micronta: .395K (.14% high)

I don't know, but I think my Micronta is not too shabby.

It's not, it's a perfectly fine meter for general use.

Dave :)
 
A

Anthony Fremont

David said:
Yeah, or they might simply follow the "customer is always right"
policy, and just wanted to keep you happy.

They certainly didn't make me unhappy by that. Now Mouser on the other
hand........... e-mail to tech support at 11:00am yesterday, not a peep. I
guess they aren't much on e-mail. I'm going to go ahead and keep it for
now, at least until I have reason to suspect that something really is wrong
with it.
I'm not really a huge fan of True RMS meters, they are slow and
generally not as accurate as normal meters. If you *know* you have a
sinusoidal signal then a normal meter is better.
Switchable true RMS would be nice!

Sure does give really slow updates (seconds between each one). I'm glad I
don't have to use it unless I need it.
It's not, it's a perfectly fine meter for general use.

I looked around some of the cal sites and in some places it's one of the
very few radio shack models they will even look at. This seems to be an
almost exact copy of it:
http://www.tequipment.net/EZDigitalKD3220.asp. The pc board has KD-3200C
stamped on it. Awful cheap now compared to when I bought mine.

I sure wish my scope would hurry up and get here. :)
 
D

David L. Jones

They certainly didn't make me unhappy by that. Now Mouser on the other
hand........... e-mail to tech support at 11:00am yesterday, not a peep. I
guess they aren't much on e-mail. I'm going to go ahead and keep it for
now, at least until I have reason to suspect that something really is wrong
with it.





Sure does give really slow updates (seconds between each one). I'm glad I
don't have to use it unless I need it.



I looked around some of the cal sites and in some places it's one of the
very few radio shack models they will even look at. This seems to be an
almost exact copy of it:http://www.tequipment.net/EZDigitalKD3220.asp. The pc board has KD-3200C
stamped on it. Awful cheap now compared to when I bought mine.

I like that it uses the same jack for V/ohms and uA/mA, no lead
swapping.

What's the build quality of the Extech like inside?
Photos?
I sure wish my scope would hurry up and get here. :)

Was that the 1000 series Rigol ?

Dave :)
 